Hello
I recently went to buy an RF sweep generator and ended up bringing home a
TS-118A/AP (Bird) RF wattmeter, as one does! I have some questions:-
1. Given that the RF power sensors for the various bands/power ratings are
added external to the meter with only the coax connecting through to the
unit,how does it work? I assume the sensors are thermocouple types but
there's no separate DC circuit and I can only see a narrow constriction in
the sensor with the centre core being a very fine wire.
2. I'm missing the calibration/multiplication factor graph for the MX-1784/U
sensor, would anyone have one they'd be prepared to scan and email for me?
Any pointers to a full manual on the net would be good also.
3. Although the unit is marked 20-1400MHz at 500W, the MX-1783/U graph shows
power levels up to 1200W at 40MHz, I guess the unit is very conservatively
rated but does anyone have true figures (not that I intend to work at that
sort of power levels).
4. Is it OK to use as a straight dummy load without the sensor, I've tested
it at 2m and it seems fine but why the 20MHz lower limit? also it doesn't
measure 50 ohms at DC so I guess it can't be a straight resistor so what's
the load?

I'd appreciate all and any information you can give about the wattmeter,
it's not something I'd have planned on buying but it looks like it'll serve
all the needs I'd ever have for a high power dummy load/wattmeter other than
for the lower HF bands.
Thanks in advance
John



  #2   Report Post  
Old February 22nd 04, 04:11 PM
Gene Tomlinson
 
Posts: n/a
Default

On Sun, 22 Feb 2004 13:35:55 -0000, "owen.home"
wrote:
I have two of these units both measure 50 ohms.
Also found the manual for it. It will be a while before I could copy
it for you.

Hello
I recently went to buy an RF sweep generator and ended up bringing home a
TS-118A/AP (Bird) RF wattmeter, as one does! I have some questions:-
1. Given that the RF power sensors for the various bands/power ratings are
added external to the meter with only the coax connecting through to the
unit,how does it work? I assume the sensors are thermocouple types but
there's no separate DC circuit and I can only see a narrow constriction in
the sensor with the centre core being a very fine wire.
2. I'm missing the calibration/multiplication factor graph for the MX-1784/U
sensor, would anyone have one they'd be prepared to scan and email for me?
Any pointers to a full manual on the net would be good also.


The MX-1784 is a tuning shunt used for safety reasons
on transmitters with capacitively coupled outputs.
3. Although the unit is marked 20-1400MHz at 500W, the MX-1783/U graph shows
power levels up to 1200W at 40MHz, I guess the unit is very conservatively
rated but does anyone have true figures (not that I intend to work at that
sort of power levels).

The manual speced power levels are.

Range 20- 1400mhz
power 20-100mhz 2-300watts
100-300mhz 2-500watts
300-750mhz 2-125watts
750-1400mhz 2-125 watts
4. Is it OK to use as a straight dummy load without the sensor, I've tested
it at 2m and it seems fine but why the 20MHz lower limit? also it doesn't
measure 50 ohms at DC so I guess it can't be a straight resistor so what's
the load?

The lower 20 mhz limit is to prevent damage to the metering circuits.
It can be used to DC if you disconnect the metering circuit.
I learned the hard way!!!!!

It is a straight 50 ohm resistor in the oil. If you measure anything
much differnet the resistor is defective.

I'd appreciate all and any information you can give about the wattmeter,
it's not something I'd have planned on buying but it looks like it'll serve
all the needs I'd ever have for a high power dummy load/wattmeter other than
for the lower HF bands.
Thanks in advance
John



  #3   Report Post  
Old February 22nd 04, 04:11 PM
Gene Tomlinson
 
Posts: n/a
Default

On Sun, 22 Feb 2004 13:35:55 -0000, "owen.home"
wrote:
I have two of these units both measure 50 ohms.
Also found the manual for it. It will be a while before I could copy
it for you.

Hello
I recently went to buy an RF sweep generator and ended up bringing home a
TS-118A/AP (Bird) RF wattmeter, as one does! I have some questions:-
1. Given that the RF power sensors for the various bands/power ratings are
added external to the meter with only the coax connecting through to the
unit,how does it work? I assume the sensors are thermocouple types but
there's no separate DC circuit and I can only see a narrow constriction in
the sensor with the centre core being a very fine wire.
2. I'm missing the calibration/multiplication factor graph for the MX-1784/U
sensor, would anyone have one they'd be prepared to scan and email for me?
Any pointers to a full manual on the net would be good also.


The MX-1784 is a tuning shunt used for safety reasons
on transmitters with capacitively coupled outputs.
3. Although the unit is marked 20-1400MHz at 500W, the MX-1783/U graph shows
power levels up to 1200W at 40MHz, I guess the unit is very conservatively
rated but does anyone have true figures (not that I intend to work at that
sort of power levels).

The manual speced power levels are.

Range 20- 1400mhz
power 20-100mhz 2-300watts
100-300mhz 2-500watts
300-750mhz 2-125watts
750-1400mhz 2-125 watts
4. Is it OK to use as a straight dummy load without the sensor, I've tested
it at 2m and it seems fine but why the 20MHz lower limit? also it doesn't
measure 50 ohms at DC so I guess it can't be a straight resistor so what's
the load?

The lower 20 mhz limit is to prevent damage to the metering circuits.
It can be used to DC if you disconnect the metering circuit.
I learned the hard way!!!!!

It is a straight 50 ohm resistor in the oil. If you measure anything
much differnet the resistor is defective.

I'd appreciate all and any information you can give about the wattmeter,
it's not something I'd have planned on buying but it looks like it'll serve
all the needs I'd ever have for a high power dummy load/wattmeter other than
for the lower HF bands.
Thanks in advance
John



  #4   Report Post  
Old February 22nd 04, 07:24 PM
owen.home
 
Posts: n/a
Default

Gene
Many thanks for the reply, I understand now - the meter is across the line
but isolated at RF with an inductor so that at DC I'm seeing the meter and
load in parallel, mine measures fine with the meter disconnected.
Thanks for the warning about damaging the meter below 20MHz and the info on
the MX-1784/U, I opened it up and can see the inductor to ground.
I'd still like to know how the sensors are made up though and I would
appreciate a copy of the manual sometime if you get a chance, I'm missing
one of the sensors - are they easily available these days?
Thanks again
John

"Gene Tomlinson" wrote in message
...
On Sun, 22 Feb 2004 13:35:55 -0000, "owen.home"
wrote:
I have two of these units both measure 50 ohms.
Also found the manual for it. It will be a while before I could copy
it for you.

Hello
I recently went to buy an RF sweep generator and ended up bringing home a
TS-118A/AP (Bird) RF wattmeter, as one does! I have some questions:-
1. Given that the RF power sensors for the various bands/power ratings

are
added external to the meter with only the coax connecting through to the
unit,how does it work? I assume the sensors are thermocouple types but
there's no separate DC circuit and I can only see a narrow constriction

in
the sensor with the centre core being a very fine wire.
2. I'm missing the calibration/multiplication factor graph for the

MX-1784/U
sensor, would anyone have one they'd be prepared to scan and email for

me?
Any pointers to a full manual on the net would be good also.


The MX-1784 is a tuning shunt used for safety reasons
on transmitters with capacitively coupled outputs.
3. Although the unit is marked 20-1400MHz at 500W, the MX-1783/U graph

shows
power levels up to 1200W at 40MHz, I guess the unit is very

conservatively
rated but does anyone have true figures (not that I intend to work at

that
sort of power levels).

The manual speced power levels are.

Range 20- 1400mhz
power 20-100mhz 2-300watts
100-300mhz 2-500watts
300-750mhz 2-125watts
750-1400mhz 2-125 watts
4. Is it OK to use as a straight dummy load without the sensor, I've

tested
it at 2m and it seems fine but why the 20MHz lower limit? also it doesn't
measure 50 ohms at DC so I guess it can't be a straight resistor so

what's
the load?

The lower 20 mhz limit is to prevent damage to the metering circuits.
It can be used to DC if you disconnect the metering circuit.
I learned the hard way!!!!!

It is a straight 50 ohm resistor in the oil. If you measure anything
much differnet the resistor is defective.

I'd appreciate all and any information you can give about the wattmeter,
it's not something I'd have planned on buying but it looks like it'll

serve
all the needs I'd ever have for a high power dummy load/wattmeter other

than
for the lower HF bands.
Thanks in advance
John





  #5   Report Post  
Old February 22nd 04, 07:24 PM
owen.home
 
Posts: n/a
Default

Gene
Many thanks for the reply, I understand now - the meter is across the line
but isolated at RF with an inductor so that at DC I'm seeing the meter and
load in parallel, mine measures fine with the meter disconnected.
Thanks for the warning about damaging the meter below 20MHz and the info on
the MX-1784/U, I opened it up and can see the inductor to ground.
I'd still like to know how the sensors are made up though and I would
appreciate a copy of the manual sometime if you get a chance, I'm missing
one of the sensors - are they easily available these days?
Thanks again
John

"Gene Tomlinson" wrote in message
...
On Sun, 22 Feb 2004 13:35:55 -0000, "owen.home"
wrote:
I have two of these units both measure 50 ohms.
Also found the manual for it. It will be a while before I could copy
it for you.

Hello
I recently went to buy an RF sweep generator and ended up bringing home a
TS-118A/AP (Bird) RF wattmeter, as one does! I have some questions:-
1. Given that the RF power sensors for the various bands/power ratings

are
added external to the meter with only the coax connecting through to the
unit,how does it work? I assume the sensors are thermocouple types but
there's no separate DC circuit and I can only see a narrow constriction

in
the sensor with the centre core being a very fine wire.
2. I'm missing the calibration/multiplication factor graph for the

MX-1784/U
sensor, would anyone have one they'd be prepared to scan and email for

me?
Any pointers to a full manual on the net would be good also.


The MX-1784 is a tuning shunt used for safety reasons
on transmitters with capacitively coupled outputs.
3. Although the unit is marked 20-1400MHz at 500W, the MX-1783/U graph

shows
power levels up to 1200W at 40MHz, I guess the unit is very

conservatively
rated but does anyone have true figures (not that I intend to work at

that
sort of power levels).

The manual speced power levels are.

Range 20- 1400mhz
power 20-100mhz 2-300watts
100-300mhz 2-500watts
300-750mhz 2-125watts
750-1400mhz 2-125 watts
4. Is it OK to use as a straight dummy load without the sensor, I've

tested
it at 2m and it seems fine but why the 20MHz lower limit? also it doesn't
measure 50 ohms at DC so I guess it can't be a straight resistor so

what's
the load?

The lower 20 mhz limit is to prevent damage to the metering circuits.
It can be used to DC if you disconnect the metering circuit.
I learned the hard way!!!!!

It is a straight 50 ohm resistor in the oil. If you measure anything
much differnet the resistor is defective.

I'd appreciate all and any information you can give about the wattmeter,
it's not something I'd have planned on buying but it looks like it'll

serve
all the needs I'd ever have for a high power dummy load/wattmeter other

than
for the lower HF bands.
Thanks in advance
John





Reply
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules

Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Why Match ? PDRUNEN Antenna 16 December 22nd 04 07:40 AM
Tx Source Impedance & Load Reflections Richard Fry Antenna 8 May 28th 04 06:29 PM
bird dummy load construction.. zindazenda Homebrew 1 February 3rd 04 10:21 PM
RF amps: tuned load in Class A? Don Pearce Homebrew 141 September 15th 03 05:02 AM
Load matching puzzle Tom Bruhns Antenna 4 August 24th 03 01:22 AM


All times are GMT +1. The time now is 12:37 PM.

Powered by vBulletin® Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Copyright ©2004-2024 RadioBanter.
The comments are property of their posters.
 

About Us

"It's about Radio"

 

Copyright © 2017