Tektronix 7603 scope behavior when readout board is missing
On 7 Mai, 04:12, msg wrote:
msg wrote:
Greetings:
I just acquired a Tek 7603 scope with 7A18 dual-trace vert. and 7B53A
timebase plugins...
snip
The display is very corrupted: odd blanking and sawtooth displays
appear with no vertical input and horizontal sweep width changes with
timebase selection in a drastic way. A vertical input signal modulates
the strange displays in a fashion that suggests the vertical section
is functional and the vertical position controls do work.
The problem was a defective C811 (14000 uf / 25v), the -18V filter
capacitor for the -15V supply.
Scope works well now.
Regards,
Michael
Sounds good to my repair eperience. Due to own experience bad/dried
filter caps in the LV pwr supply of Tek 76xx is often a problem. I had
more than 50 of these for repair, and most of these due to cap
problems.
Good luck with your fresh repaired scope, it is a nice instrument
(with and without readout option).
hth,
Andreas
|