On 7 Mai, 04:12, msg wrote:
msg wrote: Greetings: I just acquired a Tek 7603 scope with 7A18 dual-trace vert. and 7B53A timebase plugins... snip The display is very corrupted: odd blanking and sawtooth displays appear with no vertical input and horizontal sweep width changes with timebase selection in a drastic way. A vertical input signal modulates the strange displays in a fashion that suggests the vertical section is functional and the vertical position controls do work. The problem was a defective C811 (14000 uf / 25v), the -18V filter capacitor for the -15V supply. Scope works well now. Regards, Michael Sounds good to my repair eperience. Due to own experience bad/dried filter caps in the LV pwr supply of Tek 76xx is often a problem. I had more than 50 of these for repair, and most of these due to cap problems. Good luck with your fresh repaired scope, it is a nice instrument (with and without readout option). hth, Andreas |