LinkBack Thread Tools Search this Thread Display Modes
Prev Previous Post   Next Post Next
  #5   Report Post  
Old July 18th 08, 12:58 PM posted to rec.radio.amateur.homebrew
external usenet poster
 
First recorded activity by RadioBanter: Jan 2007
Posts: 58
Default Measure Signal Generator with Oscilloscope?

Michael Black wrote:
On Thu, 17 Jul 2008, Bert Hyman wrote:

() wrote in
:

I know the scope is working because its 1K hz calibration signal
looks fine.



Well, you know that the path from the calibrator through the scope is
OK, but probably nothing about the external inputs.

The only scope calibrators I've seen have been a jack or something on the
front panel of the scope, so when you touch the probe to it, you take
into consideration the scope probe, the cable to the scope and the input
circuitry. The "scope calibrator" is as useful for a quick check like
this as an potential calibrating use.


The generator dial goes from 100 KC to 30 MC. Is this too high a
frequency to be monitored with a 465?



The 465's specs were DC to 100MHz.

And as someone pointed out when he asked elsewhere, a digitally tuned
radio will be useful. Even an analog, since you can at least tell if
the generator is putting out a signal by tuning the radio and the
generator to the same frequency, and adjusting one slightly if there's
no output (to ensure one isn't off-frequency). A shortwave radio has
the advantage over the average table radio since it can tune higher in
frequency. When I had an SP-600, it was about the best piece of test
equipment I had.

Michael VE2BVW

Well you should see the signal fine on a 100MHz 465.
I have a very similar generator and I was just using it today to set up
some filters with no problems at ~7.8 Mhz, using a 60Mhz Telequipment scope.
Does that model Heath sig Gen have an output meter? Make sure the coarse
attenuator switch and fine level adjustments are fully clockwise. You
should get more than full scale on the meter when switched to Rf level.
If thats Ok then you should get around 100mV or more at the scope input.
Are you perhaps using a /100 probe, there are a few around. Most scope
probes are x1 or /10 switched though.
If the scope is indeed working and your trigger is set to auto with the
Y input at ~50Mv/div you should get plenty of signal at all of the
generators frequency range. If no display then you may have trigger
problems at higher freqencies with the scope.
Cliff Wright ZL1BDA



 
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules

Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
howto measure the freq of a received signal Caveat Lector Antenna 0 February 4th 07 09:46 PM
HP 8444A Tracking Generator, Agilent HP 182T Large Screen Oscilloscope, hardware, manuals etc. for sale Mr Mike Swap 0 September 8th 05 12:33 AM
FA: $9.99 AIRFORCE SIGNAL CORPS AN/USM-32 Oscilloscope, TUBE-TYPE RLucch2098 Equipment 0 July 13th 04 02:38 PM
FA: $9.99 AIRFORCE SIGNAL CORPS AN/USM-32 Oscilloscope, TUBE-TYPE RLucch2098 Equipment 0 July 13th 04 02:38 PM
FA: $9.99 AIRFORCE SIGNAL CORPS AN/USM-32 Oscilloscope, TUBE-TYPE RLucch2098 Swap 0 July 13th 04 02:38 PM


All times are GMT +1. The time now is 10:08 AM.

Powered by vBulletin® Copyright ©2000 - 2025, Jelsoft Enterprises Ltd.
Copyright ©2004-2025 RadioBanter.
The comments are property of their posters.
 

About Us

"It's about Radio"

 

Copyright © 2017