Steve,
The MJK unit is now giving sensible results. I changed the match from 4500 to 800 Ohms. I think the inductor Q at the higher ratio of impedance transform was not high enough. Matching to lower impedance only required a Q of around 8. Steve Nosko wrote: David, Be careful of that "Match Efficiency" number read out by the MFJ. I don't remember what they are showing, but I do remember there was a caution in their manual about just what it was and that it was a rather crude indicator. Some over simplified readout. 73, Steve, K,9.D;C'I "David" wrote in message ... Amos, The circuit is passive (R/L/C) only. I would not expect any non-linearity. I suspect the previous post may have been correct. The match was over a large ratio of impedance (50 Ohms to 4500 Ohms). I expect the small value of inductor and reasonably low Q was limiting the performance of the circuit. I tried the match again from 50 Ohms to 800 Ohms as a test and found the MJ-269 showed match efficiency of 90% with SWR dropping to 1.8. These results are much better. Regards David Amos Keag wrote: Have you considered that there may be non-linear effects in the test set up? The MFJ uses several to ten milliwatts in it's circuitry [+10 dBm]. Sensitivity is generally measured at the pico-watt [~ -130 dBm] level. AK David wrote: Hi, I have a matching circuit that matches 50 Ohms to 4500 Ohms in parallel with 2.5pF at 45MHz. The coil was measured at 147nH with a Q of 50 at 45 MHz. The circuit is a C-Tap typology with following values. Tap input to ground = 784pF Tap to hot side of coil = 92pF Par inductor = 147nH My design calculations predict Loaded Q = 26 and BW = 1.7 MHz. I used a 50p trimmer in par with 68pF cap for top of tap to fine tune. A 4k7 resistor and 2p2 cap were placed across the output to simulate the correct load. The test results show BW = 1.7 MHz and Loaded Q = 24.7 (agree with calculated values). When installed on the IF front end, the expected sensitivity is achieved when tuned. This would indicate most of the signal must be getting through or sensitivity would be effected due to mismatch. All seems to look ok but when I attach the MJF-269 Analyzer I get impedance of 280 angle 35 degrees (240 -j 179) for the impedance, I get SWR = 6.1 , return loss of 2.8dB and match efficiency of 48%. I cannot understand why there is such a discrepency between the design values and tested results (using sig gen, spectrum analyzer and power meter) compared to the MJF-269. Is there something I am doing wrong here or is it a mis-interpretation of displayed values ? Thanks for any help. -- Kind Regards David Huisman General Manager ----------------------------------------------------------------------- ORBIT COMMUNICATIONS Pty Ltd - Wireless Solutions that Work (Telemetry, Control, Monitoring, Security, HVAC ...) A.C.N. 107 441 869 Website : http://www.orbitcoms.com PO Box 4474 Lakehaven NSW 2263, AUSTRALIA Phone: 61-2-4393-3627 Fax : 61-2-4393-3685 Mobile: 61-413-715-986 |
Thread Tools | Search this Thread |
Display Modes | |
|
|
![]() |
||||
Thread | Forum | |||
FA: Tektronix 497P 100 Hz-21 GHz spectrum analyzer | Equipment | |||
Sangean Problems? | Shortwave | |||
problems reading W3UHF's icomclassic site, too much 'script' | Boatanchors |